[SEN0169] Voltage values from pH Meter v1.1 seem off
![userHead](https://dfimg.dfrobot.com/dfPortrait2.png)
I recently purchased the SEN0169 that has the pH meter v1.1 board. When I connect the pH sensor to the PCB and power it with 5v, the voltage readings from the Output pin (A) seem to be off as I can't calculate proper pH values.
Here are the voltage values I'm getting using pH buffer solutions
Here are the voltage readings when the BNC connector is shorted or when no sensor is connected to it
Any help would be greatly appreciated
Thank you in advance,
Greg
pH Buffer Voltage Reading
4.0 1.77v
6.86 1.993v
9.18 2.17v
Is the board bad? How can I check the pH sensor directly?Shorted BNC connector 2.00v
Nothing connected to BNC connector 4.32v
Greg
2019-01-04 15:56:43 Using the equation in my code the pH values are correct now.
What confused me was I tried the Arduino sample code posted at https://www.dfrobot.com/wiki/index.php/ ... KU:SEN0169 and was getting incorrect values.
Once I replaced:
Greg
greg.diehl.gtd
Once I replaced:
With this line:
pHValue = 3.5*voltage+Offset;
The pH values appeared properly.pHValue = (((12.95 * voltage) + (-18.92)) + Offset);
Greg
![userHeadPic](https://dfimg.dfrobot.com/dfPortrait2.png)
2019-01-04 11:23:22 The voltage you got seems OK with the buffer pH, please calibrate the sensor and output pH value to see if they are good to use.
robert.chen
2019-01-04 09:05:05 Given the pH sensor values should be linear, I plotted two values and came up with this equation
y=mx + b
y = 12.95x + (-18.92)
The above equation when supplied with the voltages (x) I mentioned previously will output the pH values (y) I'm expecting.
I'm going to put it in code and retry with the sensor.
Greg
greg.diehl.gtd
y=mx + b
Greg