Four-point Calibration for Gravity Analog Dissolved Oxygen Sensor?
Hi all!
I'd like to use a four-point calibration method for the oxygen meter instead of a two-point calibration to increase the accuracy of the measurements. I already have my measurements at 4 different temperatures but I'm not sure how to modify the formula in the code. Here is all the code before the setup. The formula I want to modify is the last bit of code. Thanks!
#include <Arduino.h>
#define DO_PIN A1
#define VREF 5000 //VREF (mv)
#define ADC_RES 1024 //ADC Resolution
//Single-point calibration Mode=0
//Two-point calibration Mode=1
#define TWO_POINT_CALIBRATION 0
#define READ_TEMP (25) //Current water temperature ℃, Or temperature sensor function
//Single point calibration needs to be filled CAL1_V and CAL1_T
#define CAL1_V (1600) //mv
#define CAL1_T (25) //℃
//Two-point calibration needs to be filled CAL2_V and CAL2_T
//CAL1 High temperature point, CAL2 Low temperature point
#define CAL2_V (1300) //mv
#define CAL2_T (15) //℃
const uint16_t DO_Table[41] = {
14460, 14220, 13820, 13440, 13090, 12740, 12420, 12110, 11810, 11530,
11260, 11010, 10770, 10530, 10300, 10080, 9860, 9660, 9460, 9270,
9080, 8900, 8730, 8570, 8410, 8250, 8110, 7960, 7820, 7690,
7560, 7430, 7300, 7180, 7070, 6950, 6840, 6730, 6630, 6530, 6410};
uint8_t Temperaturet;
uint16_t ADC_Raw;
uint16_t ADC_Voltage;
uint16_t DO;
int16_t readDO(uint32_t voltage_mv, uint8_t temperature_c)
#if TWO_POINT_CALIBRATION == 0
uint16_t V_saturation = (uint32_t)CAL1_V + (uint32_t)35 * temperature_c - (uint32_t)CAL1_T * 35;
return (voltage_mv * DO_Table[temperature_c] / V_saturation);
#else
uint16_t V_saturation = (int16_t)((int8_t)temperature_c - CAL2_T) * ((uint16_t)CAL1_V - CAL2_V) / ((uint8_t)CAL1_T - CAL2_T) + CAL2_V;
#endif