DFRobot Gravity: analog electrical conductivity meter V2
Good morning. A couple of days ago we received a sensor from your house (DFR0300) and when it was time to start calibrating it with the solutions indicated in the instruction manual (1413us/cm & 12.88ms/cm ) the values are not consistent. Following the steps indicated to carry out the calibration, when we introduce the probe into the 1413 us/cm solution, the indicated value is 12.88 ms/cm (approximate). On the other hand, when introducing the probe in the 12.88 ms/cm solution, the value shown is 36.50 ms/cm (approximate). All these results have been the result of the steps indicated to follow. Any help is appreciated
Hi!
what board are you using? If you choose the Arduino UNO or Mega2560, pls upload this code to erase the contents in the specified position in EEPROM. But this EEPROM code can't work well in ESP32, you can try the ADS1115 module & ESP32 to get more accurate data.
jenna