General Gravity

Analog Electrical Conductivity Sensor

userHead Sandi Pamungkas 2023-11-03 11:03:53 364 Views2 Replies

hello
Here I am trying to use an EC sensor, I have calibrated the condition of the sensor and I continuously dip the sensor in salt water. The readings became chaotic after 1 week after I dipped the sensor, is this normal or what?

2024-02-28 04:53:24

Hi,  Could you provide us the schematics of this board? the V1 seems different for adequations

userHeadPic Luis.Correa6446
2023-11-03 14:30:58

Hello,

 

Are you currently using the DFR0300 Analog Electrical Conductivity Sensor? Link: https://www.dfrobot.com/product-1123.html 

 

On the product page, there is a warning that this sensor can not be used in liquid for a long time.

 

Please try to clean and dry the probe, wait a while and reconnect the probe to check if the readings have improved. 

 

Best regards,

userHeadPic xingzhao.zhu